Sterilisation and Disinfection — Methods, Parameters and Spaulding Classification

Written & medically reviewed by the Kinase Medical Team · Last reviewed

Quick Answer

Sterilisation destroys all microorganisms including bacterial spores; disinfection removes most but not all. Steam (autoclave) is the standard method for heat-stable items, dry heat suits metal and sharp instruments, and ethylene oxide suits heat-sensitive items. Spaulding assigns critical items to sterilisation, semicritical items to high-level disinfection and noncritical items to low-level disinfection.

What is the difference between sterilisation and disinfection?

Sterilisation is the complete destruction of all microorganisms, including bacterial spores. Disinfection eliminates most pathogenic microorganisms from inanimate objects but does not necessarily kill spores. Antiseptics are germicides applied to living tissue and skin; disinfectants are used only on inanimate objects because they can injure tissue.

The CDC guideline defines three levels of disinfection by what they leave alive. High-level disinfection kills all microorganisms except large numbers of bacterial spores. Intermediate-level disinfectants may kill mycobacteria, vegetative bacteria, most viruses and most fungi but do not necessarily kill spores. Low-level disinfectants kill most vegetative bacteria, some fungi and some viruses within about 10 minutes.

Sterilization and DisinfectionLecture on physical and chemical methods of sterilisation and disinfection in the microbiology laboratory.Video: Clinical Lab Science Videos · 12:44 · Watch on YouTube · Loads from YouTube (privacy-enhanced mode) only when you press play.
Levels of microbial kill
LevelKillsDoes not reliably kill
SterilisationAll microorganisms including bacterial sporesNothing (sterile)
High-level disinfectionAll microorganisms except large numbers of bacterial sporesLarge numbers of spores
Intermediate-level disinfectionMycobacteria, vegetative bacteria, most viruses, most fungiNot necessarily spores
Low-level disinfectionMost vegetative bacteria, some fungi, some viruses (about 10 minutes or less)Spores, mycobacteria, many non-lipid viruses

How does the Spaulding classification decide the method?

Spaulding classification assigns the required level of processing to a medical item according to the infection risk of its intended use. It is a favourite exam table because each category pairs an example item with a minimum method.

Spaulding classification (CDC guideline)
CategoryContactMinimum processingExamples
CriticalEnters sterile tissue or the vascular systemSterilisationSurgical instruments, cardiac and urinary catheters, implants, ultrasound probes used in sterile body cavities
SemicriticalMucous membranes or non-intact skinHigh-level disinfection (minimum)Respiratory therapy and anaesthesia equipment, some endoscopes, laryngoscope blades, cystoscopes, diaphragm fitting rings
NoncriticalIntact skin onlyLow-level disinfection or cleaningBedpans, blood pressure cuffs, crutches, computers

The logic follows the body barrier. Intact mucous membranes of the lung and gut resist common bacterial spores but are susceptible to other bacteria, mycobacteria and viruses, so semicritical items must be free of all organisms though a few spores are acceptable. Intact skin is an effective barrier, so the sterility of items touching only intact skin is "not critical".

How does steam sterilisation (autoclave) work?

Moist heat under pressure is the most widely used and most dependable method for heat- and moisture-stable items such as surgical instruments and linen packs. Pressure is only a means of reaching the high steam temperatures needed; the lethal agent is saturated steam (dryness fraction at least 97 per cent), which condenses on the item and releases latent heat that destroys cell walls, enzymes and structural proteins.

Steam sterilisation parameters
SourceTemperaturePressureExposure
CDC guideline, gravity displacement, wrapped instruments121 °C (250 °F)—30 minutes
CDC guideline, prevacuum, wrapped instruments132 °C (270 °F)—4 minutes
Review in Indian Journal of Ophthalmology (2026)121 °C15 psi21 minutes
Same review134 °C21 psi5 minutes
Two stainless steel vertical pressure-cooker type autoclaves with pressure gauges and clamped lids.
Vertical laboratory autoclaves: the clamped lid and pressure gauge reflect that steam is generated under pressure.Image: Frankincense Diala, CC0

Two sterilizer designs appear in exams. In a gravity displacement autoclave, steam enters at the top or sides and, being lighter than air, pushes air downwards and out through a drain; penetration of porous loads is slower. A prevacuum (high-speed) sterilizer has a vacuum pump that removes air before steam enters, giving almost instant steam penetration even into porous loads and permitting a shorter, hotter cycle.

Cutaway illustration of a rectangular hospital autoclave showing a jacketed chamber and loaded shelves.
A hospital steam sterilizer: the chamber is surrounded by a steam jacket and loaded in racks so steam reaches every surface.Image: Priorclave North America, CC BY-SA 4.0

The Bowie-Dick test checks the air-removal system of a prevacuum sterilizer. A test pack of folded clean cotton towels with a chemical indicator sheet is placed in the front, bottom part of an otherwise empty chamber, near the door and over the drain, and run at 134 °C for 3.5 minutes, each day before the first load. A uniform colour change means adequate air removal; a spot indicates trapped air.

What are the hot air oven (dry heat) parameters?

Dry heat kills mainly by oxidation of cell constituents. It penetrates materials and does not corrode metal or sharp instruments, which makes it suitable for metal and sharp instruments. Its disadvantages are slow heat penetration, a time-consuming cycle, and high temperatures that are unsuitable for most materials.

Hot air sterilisation time-temperature relationships (CDC)
TemperatureHolding time
170 °C (340 °F)60 minutes
160 °C (320 °F)120 minutes (2 hours)
150 °C (300 °F)150 minutes
Front view of a single-door laboratory hot air oven with a digital temperature controller on top.
A laboratory hot air oven: an electrically heated insulated chamber with a thermostat-controlled temperature display.Image: Srini297, CC BY-SA 4.0

There are two types. The static-air (oven-type) sterilizer heats by coils at the bottom, with hot air rising by gravity convection; it heats slowly and is less uniform. The forced-air (mechanical convection) sterilizer uses a motor-driven blower to circulate hot air at high speed, giving faster, more uniform energy transfer.

How are heat-sensitive items sterilised?

Ethylene oxide (ETO) is a colourless, flammable and explosive gas used for heat- or moisture-sensitive medical equipment. Four parameters govern its efficacy: gas concentration, temperature, relative humidity and exposure time. Its drawbacks are the lengthy cycle and aeration time, cost, and hazards to staff and patients; it is infrequently used for flexible endoscopes for this reason.

Ethylene oxide operating ranges (CDC)
ParameterRange
Gas concentration450 to 1200 mg/L
Temperature37 to 63 °C
Relative humidity40 to 80 per cent
Exposure time1 to 6 hours

Liquid chemical sterilants include 2.4 per cent or higher glutaraldehyde, 7.5 per cent stabilised hydrogen peroxide and 0.2 per cent peracetic acid; indicated sterilisation contact times range from 3 to 12 hours. They work only if cleaning precedes treatment and concentration, contact time, temperature and pH are respected. The same agents at shorter contact times serve as high-level disinfectants: for example, ortho-phthalaldehyde (0.55 per cent) has a 12-minute high-level disinfection claim.

Which disinfectants are used, and what do they miss?

Common disinfectants and key points
AgentKey facts
Alcohols (ethyl, isopropyl)Rapidly bactericidal, tuberculocidal, fungicidal and virucidal but not sporicidal; activity falls sharply below 50 per cent concentration; best at 60 to 90 per cent in water. No alcohol-based product is a cleared sterilant or high-level disinfectant.
Sodium hypochloriteChlorine-releasing agent used for surfaces contaminated with HIV or HBV; sporicidal, bactericidal and virucidal. A 1 in 100 dilution (500 to 600 ppm free chlorine) is used for blood spills.
Glutaraldehyde, ortho-phthalaldehydeHigh-level disinfectants for endoscopes; rinse thoroughly to avoid colitis after colonoscopy.
FormaldehydeBactericidal, virucidal and sporicidal but slower than glutaraldehyde; carcinogenic, so protective equipment is essential.
Povidone-iodineRapid antiseptic; staining and skin irritation; less effective against fungi and spores than chlorine agents.
ChlorhexidineSkin antiseptic with broad activity; mycobacteria are highly resistant to it.

Microorganisms differ in innate resistance. The CDC orders them from most to least resistant to chemical germicides: prions, bacterial spores, coccidia such as Cryptosporidium, mycobacteria, non-lipid or small viruses, fungi, vegetative bacteria, and finally lipid or medium-size viruses such as herpes and HIV.

How is sterilisation monitored — physical, chemical and biological indicators?

A sterilisation process is monitored with mechanical (physical) readings of time, temperature and pressure, with chemical indicators, and with biological indicators. Only biological indicators measure killing power directly, because they use the most resistant organisms, bacterial spores.

Biological indicator organisms (CDC)
ProcessIndicator sporeIncubation
Steam, hydrogen peroxide gas plasma, liquid peracetic acidGeobacillus stearothermophilus (about 10⁵ spores)55 to 60 °C
Dry heat and ethylene oxideBacillus atrophaeus (about 10⁶ spores)35 to 37 °C

Chemical indicators are heat- or chemically sensitive inks that change colour when a parameter is reached. They are fixed on the outside of each pack to show it has been through a cycle, and preferably also inside to show sterilant penetration. They do not prove sterility, can mislead at marginal exposure times, and should be used alongside, not instead of, biological indicators. Chemical liquid sterilants generally cannot be monitored with a biological indicator.

What are the common exam traps in sterilisation and disinfection?

  • Pressure is not the killer. It raises steam temperature; saturated steam at 121 °C is what sterilises.
  • Alcohol, 70 per cent included, does not sterilise because it is not sporicidal; it is a disinfectant and skin antiseptic.
  • Laryngoscope blades and endoscopes are semicritical, so high-level disinfection is the minimum, not sterilisation.
  • Dry heat needs longer and hotter than steam because heat penetrates slowly: 160 °C for 2 hours versus 121 °C for about 30 minutes in a gravity autoclave.
  • The most resistant agents to germicides are prions, followed by bacterial spores.
  • Match the spore to the method: steam = G. stearothermophilus; dry heat and ETO = B. atrophaeus.
  • Chemical indicator colour change on a pack proves exposure, not sterility.

Frequently asked questions

What is the autoclave temperature and pressure for sterilisation?
A common steam cycle is 121 °C at 15 psi, with the CDC recognising 30 minutes for wrapped supplies in a gravity displacement sterilizer. Prevacuum cycles use 132 to 134 °C for roughly 4 to 5 minutes. Exact times depend on load type, wrapping and sterilizer design, so read the question for the cycle described.
What temperature and time are used in a hot air oven?
The usual time-temperature pairs for dry heat are 170 °C for 60 minutes, 160 °C for 120 minutes and 150 °C for 150 minutes. The most quoted exam answer is 160 °C for 2 hours. Dry heat suits metal and sharp instruments but is unsuitable for most materials.
Which organism is used to test an autoclave?
Geobacillus stearothermophilus spores are the biological indicator for steam sterilisation, incubated at 55 to 60 °C. Bacillus atrophaeus spores monitor dry heat and ethylene oxide, incubated at 35 to 37 °C. Biological indicators are the only indicators that directly measure killing power, unlike chemical indicators or readings of time and temperature.
How does the Spaulding classification apply to laryngoscope blades?
Laryngoscope blades touch mucous membranes, so they are semicritical items. The minimum requirement is high-level disinfection rather than sterilisation. Surgical instruments and implants are critical and need sterilisation; blood pressure cuffs and bedpans are noncritical and need low-level disinfection or cleaning.
Is 70 per cent alcohol a sterilising agent?
No. Alcohols kill vegetative bacteria, mycobacteria, fungi and many viruses quickly but do not destroy bacterial spores, so they cannot sterilise. Activity falls sharply below about 50 per cent concentration and is best at 60 to 90 per cent in water. They are used for skin and surface disinfection.
What is the Bowie-Dick test?
It is a daily air-removal test for prevacuum steam sterilizers. A pack of folded cotton towels with a chemical indicator sheet is run in an empty chamber at 134 °C for 3.5 minutes before the first load. A uniform colour change passes; a spot of unchanged indicator shows trapped air and the sterilizer should not be used.
Which method sterilises heat-sensitive critical items?
Ethylene oxide gas and hydrogen peroxide gas plasma are the low-temperature options named for heat-sensitive critical items; liquid chemical sterilants such as 2.4 per cent glutaraldehyde are used when other methods are unsuitable. Ethylene oxide operates at 37 to 63 °C with 40 to 80 per cent humidity for 1 to 6 hours.

Sources

  1. CDC — Guideline for Disinfection and Sterilization in Healthcare Facilities (2008, updated 2024)
  2. PMC — Steam sterilization: Review of autoclaves, validation, instrument packing, and sterility failures (Indian J Ophthalmol, 2026)
  3. StatPearls — Disinfectants

For exam preparation and education only — not a substitute for clinical judgement or local guidelines. How we write and review these pages: editorial policy.

Revise Sterilisation and Disinfection with questions

Kinase: NEET-PG & INICET has previous-year papers, a subject-wise QBank and Grand Tests with explanations — on Android, iOS and the web.