What is the difference between sterilisation and disinfection?
Sterilisation is the complete destruction of all microorganisms, including bacterial spores. Disinfection eliminates most pathogenic microorganisms from inanimate objects but does not necessarily kill spores. Antiseptics are germicides applied to living tissue and skin; disinfectants are used only on inanimate objects because they can injure tissue.
The CDC guideline defines three levels of disinfection by what they leave alive. High-level disinfection kills all microorganisms except large numbers of bacterial spores. Intermediate-level disinfectants may kill mycobacteria, vegetative bacteria, most viruses and most fungi but do not necessarily kill spores. Low-level disinfectants kill most vegetative bacteria, some fungi and some viruses within about 10 minutes.
| Level | Kills | Does not reliably kill |
|---|---|---|
| Sterilisation | All microorganisms including bacterial spores | Nothing (sterile) |
| High-level disinfection | All microorganisms except large numbers of bacterial spores | Large numbers of spores |
| Intermediate-level disinfection | Mycobacteria, vegetative bacteria, most viruses, most fungi | Not necessarily spores |
| Low-level disinfection | Most vegetative bacteria, some fungi, some viruses (about 10 minutes or less) | Spores, mycobacteria, many non-lipid viruses |
How does the Spaulding classification decide the method?
Spaulding classification assigns the required level of processing to a medical item according to the infection risk of its intended use. It is a favourite exam table because each category pairs an example item with a minimum method.
| Category | Contact | Minimum processing | Examples |
|---|---|---|---|
| Critical | Enters sterile tissue or the vascular system | Sterilisation | Surgical instruments, cardiac and urinary catheters, implants, ultrasound probes used in sterile body cavities |
| Semicritical | Mucous membranes or non-intact skin | High-level disinfection (minimum) | Respiratory therapy and anaesthesia equipment, some endoscopes, laryngoscope blades, cystoscopes, diaphragm fitting rings |
| Noncritical | Intact skin only | Low-level disinfection or cleaning | Bedpans, blood pressure cuffs, crutches, computers |
The logic follows the body barrier. Intact mucous membranes of the lung and gut resist common bacterial spores but are susceptible to other bacteria, mycobacteria and viruses, so semicritical items must be free of all organisms though a few spores are acceptable. Intact skin is an effective barrier, so the sterility of items touching only intact skin is "not critical".
How does steam sterilisation (autoclave) work?
Moist heat under pressure is the most widely used and most dependable method for heat- and moisture-stable items such as surgical instruments and linen packs. Pressure is only a means of reaching the high steam temperatures needed; the lethal agent is saturated steam (dryness fraction at least 97 per cent), which condenses on the item and releases latent heat that destroys cell walls, enzymes and structural proteins.
| Source | Temperature | Pressure | Exposure |
|---|---|---|---|
| CDC guideline, gravity displacement, wrapped instruments | 121 °C (250 °F) | — | 30 minutes |
| CDC guideline, prevacuum, wrapped instruments | 132 °C (270 °F) | — | 4 minutes |
| Review in Indian Journal of Ophthalmology (2026) | 121 °C | 15 psi | 21 minutes |
| Same review | 134 °C | 21 psi | 5 minutes |

Two sterilizer designs appear in exams. In a gravity displacement autoclave, steam enters at the top or sides and, being lighter than air, pushes air downwards and out through a drain; penetration of porous loads is slower. A prevacuum (high-speed) sterilizer has a vacuum pump that removes air before steam enters, giving almost instant steam penetration even into porous loads and permitting a shorter, hotter cycle.

The Bowie-Dick test checks the air-removal system of a prevacuum sterilizer. A test pack of folded clean cotton towels with a chemical indicator sheet is placed in the front, bottom part of an otherwise empty chamber, near the door and over the drain, and run at 134 °C for 3.5 minutes, each day before the first load. A uniform colour change means adequate air removal; a spot indicates trapped air.
What are the hot air oven (dry heat) parameters?
Dry heat kills mainly by oxidation of cell constituents. It penetrates materials and does not corrode metal or sharp instruments, which makes it suitable for metal and sharp instruments. Its disadvantages are slow heat penetration, a time-consuming cycle, and high temperatures that are unsuitable for most materials.
| Temperature | Holding time |
|---|---|
| 170 °C (340 °F) | 60 minutes |
| 160 °C (320 °F) | 120 minutes (2 hours) |
| 150 °C (300 °F) | 150 minutes |

There are two types. The static-air (oven-type) sterilizer heats by coils at the bottom, with hot air rising by gravity convection; it heats slowly and is less uniform. The forced-air (mechanical convection) sterilizer uses a motor-driven blower to circulate hot air at high speed, giving faster, more uniform energy transfer.
How are heat-sensitive items sterilised?
Ethylene oxide (ETO) is a colourless, flammable and explosive gas used for heat- or moisture-sensitive medical equipment. Four parameters govern its efficacy: gas concentration, temperature, relative humidity and exposure time. Its drawbacks are the lengthy cycle and aeration time, cost, and hazards to staff and patients; it is infrequently used for flexible endoscopes for this reason.
| Parameter | Range |
|---|---|
| Gas concentration | 450 to 1200 mg/L |
| Temperature | 37 to 63 °C |
| Relative humidity | 40 to 80 per cent |
| Exposure time | 1 to 6 hours |
Liquid chemical sterilants include 2.4 per cent or higher glutaraldehyde, 7.5 per cent stabilised hydrogen peroxide and 0.2 per cent peracetic acid; indicated sterilisation contact times range from 3 to 12 hours. They work only if cleaning precedes treatment and concentration, contact time, temperature and pH are respected. The same agents at shorter contact times serve as high-level disinfectants: for example, ortho-phthalaldehyde (0.55 per cent) has a 12-minute high-level disinfection claim.
Which disinfectants are used, and what do they miss?
| Agent | Key facts |
|---|---|
| Alcohols (ethyl, isopropyl) | Rapidly bactericidal, tuberculocidal, fungicidal and virucidal but not sporicidal; activity falls sharply below 50 per cent concentration; best at 60 to 90 per cent in water. No alcohol-based product is a cleared sterilant or high-level disinfectant. |
| Sodium hypochlorite | Chlorine-releasing agent used for surfaces contaminated with HIV or HBV; sporicidal, bactericidal and virucidal. A 1 in 100 dilution (500 to 600 ppm free chlorine) is used for blood spills. |
| Glutaraldehyde, ortho-phthalaldehyde | High-level disinfectants for endoscopes; rinse thoroughly to avoid colitis after colonoscopy. |
| Formaldehyde | Bactericidal, virucidal and sporicidal but slower than glutaraldehyde; carcinogenic, so protective equipment is essential. |
| Povidone-iodine | Rapid antiseptic; staining and skin irritation; less effective against fungi and spores than chlorine agents. |
| Chlorhexidine | Skin antiseptic with broad activity; mycobacteria are highly resistant to it. |
Microorganisms differ in innate resistance. The CDC orders them from most to least resistant to chemical germicides: prions, bacterial spores, coccidia such as Cryptosporidium, mycobacteria, non-lipid or small viruses, fungi, vegetative bacteria, and finally lipid or medium-size viruses such as herpes and HIV.
How is sterilisation monitored — physical, chemical and biological indicators?
A sterilisation process is monitored with mechanical (physical) readings of time, temperature and pressure, with chemical indicators, and with biological indicators. Only biological indicators measure killing power directly, because they use the most resistant organisms, bacterial spores.
| Process | Indicator spore | Incubation |
|---|---|---|
| Steam, hydrogen peroxide gas plasma, liquid peracetic acid | Geobacillus stearothermophilus (about 10⁵ spores) | 55 to 60 °C |
| Dry heat and ethylene oxide | Bacillus atrophaeus (about 10⁶ spores) | 35 to 37 °C |
Chemical indicators are heat- or chemically sensitive inks that change colour when a parameter is reached. They are fixed on the outside of each pack to show it has been through a cycle, and preferably also inside to show sterilant penetration. They do not prove sterility, can mislead at marginal exposure times, and should be used alongside, not instead of, biological indicators. Chemical liquid sterilants generally cannot be monitored with a biological indicator.
What are the common exam traps in sterilisation and disinfection?
- Pressure is not the killer. It raises steam temperature; saturated steam at 121 °C is what sterilises.
- Alcohol, 70 per cent included, does not sterilise because it is not sporicidal; it is a disinfectant and skin antiseptic.
- Laryngoscope blades and endoscopes are semicritical, so high-level disinfection is the minimum, not sterilisation.
- Dry heat needs longer and hotter than steam because heat penetrates slowly: 160 °C for 2 hours versus 121 °C for about 30 minutes in a gravity autoclave.
- The most resistant agents to germicides are prions, followed by bacterial spores.
- Match the spore to the method: steam = G. stearothermophilus; dry heat and ETO = B. atrophaeus.
- Chemical indicator colour change on a pack proves exposure, not sterility.